WORKS!
On Mon, Jun 4, 2012 at 7:32 PM, Pavel Masloff <[email protected]> wrote:
Oh, thanks Rod! I knew I had to try this, the only confusion was how to get the macro into the protocol file :) Thanks.
~Pavel
On Mon, Jun 4, 2012 at 7:23 PM, Rod Nussbaumer <[email protected]> wrote:
If I understand your question correctly, I think it may be solved by using one StreamDevice protocol to do multiple set and query commands. In your case, setting the probe attenuation record has the side effect of altering the vertical sensitivity of the respective scope channel. You can use one protocol to write multiple commands, and to process the results into multiple records. See the StreamDevice docs for 'redirection',
<http://epics.web.psi.ch/software/streamdevice/doc/formats.html#redirection>
Using this method, you can set the attenuation on the scope, and at the same time update the scale record in you database to keep everything in sync.
setY {
ReplyTimeout = 5000;
out 'CH\$1:SCA %g';
#in "CH\$1:SCA?";
@init {
out "CH\$1:SCA?";
in ":CH\$1:SCALE %g";
probe {
out 'CH\$1:PRO %g';
out 'CH\$1:SCALE?";
in '%(\$2)g';
@init {
out 'CH\$1:PRO?';
in ':CH\$1:PROBE %g';
# New argument to the 'probe' protocol is name of accordant
# scope vertical scale record.
record(ao, "$(P)$(R)probe_ch1") {
field(DESC, "Probe voltage attenuation 1")
field(DTYP, "stream")
field(OUT, "@devTPS20xx.proto probe(1,$(P)$(R)setY1) $(PORT) $(A)")
--- rod.
Pavel Masloff wrote:
Hi Rod,
You mentioned you are using StreamDevice to talk to your Tek scopes. I
am having difficulties with the Probe Attenuator setting. So you
basically change it say from 1 to 10, and the scope automatically
changes the voltage on a channel from say 1V to 10V/DIV.
I don't want StreamDevice to poll the scope once a second (or even worse
I/O interrupt) to get this change (wanna leave it Passive), so I can
change my Attenuator PV (Passive record) it then forwards the
CHx_Voltage to ask the scope the current Y-axis resolution.
The problem is that the AO record can be either closed-loop (via db
links) or open-loop (via CA).
I don't know what to do.
In other words:
On the screenshot is my HMI
(http://imageshack.us/photo/my-images/337/scope.jpg/). So when I change
the Attenuator_1 setting (probe_ch1) from 1 to 10, my CH_1 voltage
(setY1) remains the same = 0,2 V, when in fact on the scope it is 2 V.
Is there a way my setY1 could update its value, but not process?
record(ao, "$(P)$(R)setY1")
field(DESC, "Set VOLTS/DIV channel 1")
field(OUT, "@devTPS20xx.proto setY(1) $(PORT) $(A)")
field(EGU, "V")
field(OUT, "@devTPS20xx.proto probe(1) $(PORT) $(A)")
setY {
out 'CH\$1:SCA %g';
#in "CH\$1:SCA?";
@init {
out "CH\$1:SCA?";
probe {
out 'CH\$1:PRO %g';
@init {
out 'CH\$1:PRO?';
#getY {
# out 'CH\$1:SCA?';
# in "%g";
Hope everything is understandable.
Best regards,
Pavel Maslov, MS
Controls Engineer at Pulsed power Lab
Efremov Institute for Electro-Physical Apparatus
St. Petersburg, Russia
Mobile: +7 (951) 672 22 19
Landline: +7 (812) 461 01 01
-- Best regards,
Pavel Maslov, MS Controls Engineer at Pulsed power Lab Efremov Institute for Electro-Physical Apparatus
-- Best regards,
- [StreamDevice] ? Pavel Masloff
- Re: [StreamDevice] ? Pavel Masloff
counter of clients for specific PV in IOC Yingbing Yan
ASYN 4-19 createParam return status Hinko Kocevar
<2012>
Quiz for C experts Andrew Johnson
<2012>
ANJ, 18 Nov 2013 |
Home
News
About
Base
Modules
Extensions
Distributions
Download
Search
EPICS V4
IRMIS
Talk
Bugs
Documents
Links
Licensing